Eddy current testing using ms5800

  • Eddy current testing using ms5800

    Posted by caleb-charles on 28/11/2021 at 6:41 pm

    Dear friends,
    In Olympus MS5800 using multi view software,
    Is it any possible way to change the calibration point value for the ID GROOVE after the inspection?
    Becoz I have stack with this issue
    By creating the setup wizard, I have given 40 deg for idgroove instead of 10deg, I have facing inadequete phase ref in analysis plane, it all because my careless, by creating a new setup, I can make reinspection, but here I want to know… Is it anyway for resolve this kind issue?

    william replied 2 years, 11 months ago 2 Members · 1 Reply
  • 1 Reply
  • Eddy Current (ECT)

    william

    Member
    28/11/2021 at 6:56 pm

    Yes you can do, but it depends. If you had selected correct frequency and you have saved the calibration tube data, then you can recalibrate as earlier.

    In case, if you dont have the cal tube data, take a cal tube data at the same frequency at which you have inspected the exchanger and calibrate it. this is not technically right method but it will full-fill your requirement. We all know, that ECT is a comparative method. So, here you will get better reference.