Olympus MS5800 MFL calibration

  • Olympus MS5800 MFL calibration

    Posted by leo on 27/11/2021 at 7:56 pm

    Dear friends,
    I am working on Olympus MS5800 equipment. In that MFL calibration part absolute coil should give 6v for 60% WT OD groove, but I am getting only 4v. How to achieve 6v?
    I tried with gain adjustment. Also the differential and trial coil not giving 1v.

    caleb-charles replied 3 years ago 2 Members · 1 Reply
  • 1 Reply
  • Magnetic Flux Leakage (MFL)

    caleb-charles

    Member
    28/11/2021 at 4:29 pm

    In order to get the values you set in Setup , you need to chech the Channel option in the System Calibration window (it should show only HOLE,OD GR60 ). After recording a signal for each one you hit Callibrate and the Differential and absolute refferences will be adjusted at the value in the Setup.