PAUT Focusing problems (Near Field Length Compensation Options)

  • Started by ketkon · 4 years ago
  • 1 reply
  • 2 voices
  • last reply 2 months ago

Posted 4 years ago

Abstract

In a previous report by the authors, the relative accuracy of Beamtool as it relates to calculations of several beam parameters was considered. Comparison of the values obtained by Beamtool, to those obtained from Civa, indicated that the traditional approximations used in Beamtool were reasonably close to what might be predicted for actual probes. Deviation from the ideal was noted to occur for the near field length of some probe configurations that would be greater than desirable for even just a simple approximation.

Real life experience as well as analytical modelling has indicated that the near zone of refracted transverse wave beams experiences a shortening as the refracted angle increases. This paper considers some of the various compensation techniques used to reduce the estimated near zone. Comparison of the various techniques is done using a variety of common probe frequencies and apertures.

Keywords: Modelling, near field, validation, ultrasound, Civa

1. Introduction

The authors have previously reported on the relative accuracy of the quantitative aspects of Beamtool as it relates to calculations for refracted angle, beam divergence and near field length for both round and rectangular shaped elements. It was demonstrated that, in general, the values obtained by Beamtool compared well with those obtained from Civa. This indicated that the traditional approximations used in Beamtool were reasonably close to what might be predicted for actual probes.

When the near field lengths were compared for a small sample of apertures and shapes that might be used in most applications, it was found that some were very close to the expected real values. However, for the larger values of near field some of the values estimated by Beamtool had deviations that might be considered more than tolerable as a simple first approximation for the purposes of technique development.

Anyone who has been required to construct a distance amplitude correction (DAC) curve using a standard probe on interchangeable refracting wedges has noted that, for larger diameter probes, the peak of the curve tends to migrate to slightly shorter soundpaths when the same probe is used on higher angle refracting wedges. This effect is conformed using analytical modelling by Civa simulation software.

Figure 1 indicates the DAC that would result for a 2MHz 20mm diameter element on a 45° and 60° refracting wedge. The black line indicates the DAC for the 45° and is seen to peak at 52mm. The DAC for the 60° wedge is represented by the red line and it peaks at 13mm.


Please download file for full study case

1 reply

2 months ago

This is a valuable contribution on an important practical problem in PAUT. The observation that near field length shortens with increasing refracted angle, and that DAC curve peaks shift to shorter sound paths on higher-angle wedges, reflects real physics that practitioners need to account for in technique development.

The key takeaway for technicians and inspectors is that standard near field approximations—whether from Beamtool or hand calculations—can overestimate the usable near field, especially for larger diameter probes and steep refraction angles. This matters because an overestimated near field can lead to false confidence in sensitivity at shorter distances, where the beam is still developing.

For practical technique work, this means:

- Construct your DAC curves empirically on the actual wedge angle you will use in service, not just on a reference angle. This captures the real near field behaviour for your specific setup.

- When designing inspection procedures, validate focal distance and sensitivity claims with your actual probe and wedge combination, particularly for angles above 45°.

- If using modelling software, compare predictions against measured data on your equipment. Tools like Civa can model this effect more accurately than traditional approximations, but field validation remains essential.

- Document which compensation technique your procedure uses (if any) so that future operators and auditors understand the basis for your setup.

The paper's comparison of compensation methods is useful; posting or linking to the full study would help members apply these techniques to their own probe configurations. Has anyone here implemented one of these compensation approaches in a production procedure? Experience notes on what worked would be valuable.

More in Phased Array

All discussions in this board

966 discussions · 392 answers

Latest forum discussions

Inspection technician working on pipework at an industrial plant

Hiring?

Post a vacancy where NDT people look first

Free for companies and individuals. Reach 5,484 inspection professionals and 1,546 resumes on file.

Sponsors of NDT Inspect
Magnaflux
Sponsor slot open